Quick poll...do you users change out the exp. channel after each patient use? If yes, do you then autoclave the parts including flow transducer before re-using? | ||
|
Member |
We do change after each use. We pateurize all parts except the transducer which is soaked in 70% ETOH for 1 hour. john esbenshade | |||
|