continuous nebulization in-line to hamilton, galileo vent for infant and pediatric setup
 Login/Join
 
<Jorie>
posted

Question:
What is the maximum external flow that the ventilator can handle, and compensate for?
Any suggestions on appropriate continuous neb setups in-line to vent, that work effectively to deliver meds?

Choices:
Found that 4L/min or less works without altering pressures.
Currently using mini-heart neb "t'd" into wet side of heater, with external flows between 3-4L/min

 
 
Reply With QuoteReport This Post
<Jorie>
posted
Contacted Hamilton company, and they do not recommend any external flow because it alters the internal flow of the circuit. 1L/min external flow = 16.6cc/sec extra flow to circuit. When external flow is added in excess, the machine may stop internal flow (circuit) delivery.
Apparently, other users of Galileo vent have run into similar issues when trying to setup continuous nebulization to vent. Hamilton is working with the company that makes the aerogen nebulizer which is similar to ultra sonic neb, but doesn't add extra flow to circuit.
 